Electrical impedance tomography captures heterogeneous lung ventilation that may be associated with ineffective inspiratory efforts

Dear editor,

Patient-ventilator asynchrony (PVA) is common in patients receiving mechanical ventilation, which is due to mismatch between neural and mechanical inspiratory time. This occurs primarily when the triggering and cycling-off of ventilatory assistance are not synchronized with the patient's inspiratory efforts. Ineffective inspiratory effort during expiration (IEE) is one of the most frequent type of PVA [1], which is associated with worse clinical outcomes [2]. Despite the extensive investigations on its mechanisms and recognition, rare notice was put on the regional ventilation distribution during IEE or other types of PVA. Electrical impedance tomography (EIT) emerges as an effective tool to monitor the regional lung ventilation [3]. In this study, we aim to describe an EIT-based method to assess heterogeneous lung ventilation, which may be associated with IEE.

Synchronized EIT and ventilator waveform were recorded in three patients with acute exacerbation of chronic obstructive pulmonary disease using PulmoVista500 and V300 (Draeger Medical, Luebeck, Germany). Patients were intubated and ventilated under assist-control mode. The relative impedance changes in ventral and dorsal regions are denoted as IV(t) and ID(t), respectively. A so-called regional intensity fraction (RIF) curve was calculated as

$${\text{RIF}}\left( t \right) = \frac{{{\text{Normalized}}\,I_{{\text{V}}}^{\prime } \left( t \right)}}{{{\text{Normalized}}\,I_{{\text{D}}}^{\prime } \left( t \right)}},$$
(1)

where I′(t) denotes the 1st order derivative of I(t). The normalization converts the I′(t) into the range between 0.1 and 1 for the feasibility of fractional calculation according to the following equation

$${\text{Norm}} \cdot I^{\prime } \left( t \right) = \frac{{\left( { I^{\prime } \left( t \right) - {\text{Min}}} \right) \times \left( {1 - 0.1} \right)}}{{{\text{Max}} - {\text{Min}}}} + 0.1$$
(2)

where \(\mathrm{Max}\) and \(\mathrm{Min}\) are the maximum and minimum value of I′(t). In order to examine the correlation between RIF and IEE events, the IEE events were visually inspected by a senior respiratory therapist with clinical experience > 10 years based on the stored ventilator waveforms.

We analyzed 6022 breath cycles in total. The clinical expert annotated 2945 IEE cycles and 3077 non-IEE cycles. In the expiratory phase, non-IEE cycles do not exhibit a local maximum (Fig. 1a), whereas typical IEE cycles show a distinct one (Fig. 1b). RIFmax indicates the peak in the expiratory phase. RIFmin denotes the local minimum between 0.3 s after the start of expiration and the position of RIFmax. Delta RIF, defined as \(\Delta {\text{RIF}} = {\text{RIF}}_{\max } - {\text{RIF}}_{\min }\), was associated with IEE. By selecting a threshold of 0.25 for ΔRIF, which was defined by searching the optimal value between 0.1 and 0.4 with a step of 0.05 for highest correlation between the annotation and the ΔRIF-based identification, all the 2945 IEE cycles were correctly identified as IEE by EIT. Furthermore, 920 cycles labeled as non-IEE were recognized as IEE by the proposed EIT-based method. A typical cycle with discrepant results by the ventilator waveforms and the EIT is given in Fig. 1c. Characteristic inspiratory flow was not found in the flow-time curve. However, the RIF curve shows distinct IEE features, with a significant difference between the local maximum and minimum in the expiratory phase.

Fig. 1
figure 1

The ventilator flow-time curves (first row), the normalized 1st order derivative of regional EIT intensity (second row), and the regional intensity fraction (RIF) curves (third row) of a non-IEE, b IEE, which coincided with the characteristic inspiratory flow observed in the flow-time curve, c IEE, which was not able to be identified in the corresponding flow-time curve

To our best knowledge, this is the first time to uncover that a particular regional lung ventilation pattern may be associated with the occurrence of IEE. The spontaneously ineffective breath results in more imbalanced ventilation distribution, characterized by higher portion in the dorsal regions. We suspected that patient’s inspiratory effort triggered the redistribution of ventilation as proposed in previous studies [4]. However, some of these efforts were not transferred to the airway opening due to e.g. obstructive airway.

In conclusion, EIT is able to characterize the imbalanced ventilation that may be associated with IEE. It has the potential to discover IEE cycles without corresponding characteristics in the ventilator waveform. The findings require further validations with more subjects to confirm whether these cycles reflect asynchronous breath using esophageal pressure or electrical diaphragm activity, and if they will compromise the ventilatory support for the patients.

Availability of data and materials

The datasets used and/or analyzed during the current study are available from the corresponding authors on reasonable request.

References

  1. Blanch L, Villagra A, Sales B, Montanya J, Lucangelo U, Luján M, García-Esquirol O, Chacón E, Estruga A, Oliva JC, et al. Asynchronies during mechanical ventilation are associated with mortality. Intensive Care Med. 2015;41(4):633–41.

    Article  Google Scholar 

  2. Vaporidi K, Babalis D, Chytas A, Lilitsis E, Kondili E, Amargianitakis V, Chouvarda I, Maglaveras N, Georgopoulos D. Clusters of ineffective efforts during mechanical ventilation: impact on outcome. Intensive Care Med. 2017;43(2):184–91.

    Article  Google Scholar 

  3. Borges JB, Suarez-Sipmann F, Bohm SH, Tusman G, Melo A, Maripuu E, Sandstrom M, Park M, Costa EL, Hedenstierna G, et al. Regional lung perfusion estimated by electrical impedance tomography in a piglet model of lung collapse. J Appl Physiol (1985). 2012;112(1):225–36.

    Article  Google Scholar 

  4. Zhao Z, Peng SY, Chang MY, Hsu YL, Frerichs I, Chang HT, Moller K. Spontaneous breathing trials after prolonged mechanical ventilation monitored by electrical impedance tomography: an observational study. Acta Anaesthesiol Scand. 2017;61(9):1166–75.

    Article  CAS  Google Scholar 

Download references

Acknowledgements

Not applicable.

Funding

The study is partially supported by National Natural Science Foundation of China (Grant Nos. 82070087, 52077216, 31870938).

Author information

Authors and Affiliations

Authors

Contributions

ZZ, HG and QP conceived and designed the study; QP and MJ performed the experiment, analyzed the data, and drafted the manuscript; HG and ZZ edited and revised the manuscript; all the authors approved the final version of manuscript. All authors read and approved the final manuscript.

Corresponding authors

Correspondence to Huiqing Ge or Zhanqi Zhao.

Ethics declarations

Ethics approval and consent to participate

The ethics review board of Sir Run Run Shaw Hospital approved the study protocol (No. 20190916-16), and written informed consent was obtained from all participants before enrollment.

Consent for publication

Not applicable.

Competing interests

Zhanqi Zhao receives a consulting fee from Dräger Medical. Other authors declare no conflict of interest.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Pan, Q., Jia, M., Ge, H. et al. Electrical impedance tomography captures heterogeneous lung ventilation that may be associated with ineffective inspiratory efforts. Crit Care 25, 303 (2021). https://0-doi-org.brum.beds.ac.uk/10.1186/s13054-021-03732-0

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://0-doi-org.brum.beds.ac.uk/10.1186/s13054-021-03732-0