- Meeting abstract
- Published:
Ventilator beyond ventilation: impact of the increase of elastic recoil pressure of a respiratory system through positive end expiratory pressure–zero positive end expiratory pressure vs manually assisted cough vs manually assisted cough adding positive end expiratory pressure 12 cmH2O on the peak expiratory flow in mechanically ventilated patients
Critical Care volume 7, Article number: P46 (2003)
Introduction
Physiotherapy has many techniques that are able to promote secretion clearance, decreasing airway obstruction in the attempt to improve ventilation homogeneity and exchanged blood gases [1–3]. These techniques aim to promote the increase of peak expiratory flow (PEF) and expiratory volume, probably through augmentation of the elastic recoil pressure produced by exsufflation forces. Our purpose was to assess the impact of the three usual modalities of bronchial hygiene on peak expiratory flow.
Design
A prospective, experimental and blinded study.
Methods
Twenty-six patients (mean age, 60 ± 13.5 years) admitted to the Hospital Português Clinical–Surgery Intensive Care Unit, requiring total support ventilation and with positive end expiratory pressure (PEEP) of 5.5 ± 1.44 cmH2O, were assessed. Patients with a history of pulmonary disease, hemodynamic instability, rib cage and/or abdominal abnormalities, scoliosis, pregnancy, obesity, cardiac pacemaker, pneumothorax, unstable thorax and PEEP higher than 10 cmH2O were excluded. PEF was measured by Navigator Graphics Monitor (Newport Medical Instruments Inc., Newport Beach, CA, USA) during PEEP–zero positive end expiratory pressure (ZEEP), manually assisted cough with baseline PEEP (MAC), mean PEEP of 5.5 ± 1.44 cmH2O, and MAC with adding PEEP of 12 cmH2O. Techniques were performed five times for each patient, with intervals of 5 min between each technique application. The Kruskal–Wallis test was used to compare the groups. The Mann–Whitney rank-sum test was used to compare the techniques two by two. Differences with P < 0.05 were considered significant.
Results
Means of the PEF variations are presented in Table 1. The comparison of MAC vs PEEP–ZEEP and MAC vs MAC with PEEP of 12 cmH2O showed statistical significance (P < 0.0002 and P < 0.01, respectively). When PEEP–ZEEP and MAC with PEEP of 12 cmH2O were compared, no statistical significant differences were noted.
Conclusion
PEEP–ZEEP and MAC with PEEP of 12 cmH2O were more efficient on the increase of peak expiratory flow than MAC with baseline PEEP. Considering that there were no statistical differences between the more efficient techniques, the relationship of risk/benefit and the institutional aspects, the choice of the technique should be powerful to consider.
References
Hardy KA: A review of airway clearance. New techniques, indications and recommendations. Respir Care 1994., 39:
Van Der Schans CP, Postma DS, Koeter GH, Rubin BK: Physiotherapy and bronquial mucus transport. Eur Respir J 1999, 13: 1477-1486.
Hess DR: The evidence for secretion clearance techniques. Respir Care 2001, 46: 1276-1292.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Duarte, A., Avena, K., Sampaio, A. et al. Ventilator beyond ventilation: impact of the increase of elastic recoil pressure of a respiratory system through positive end expiratory pressure–zero positive end expiratory pressure vs manually assisted cough vs manually assisted cough adding positive end expiratory pressure 12 cmH2O on the peak expiratory flow in mechanically ventilated patients. Crit Care 7 (Suppl 3), P46 (2003). https://0-doi-org.brum.beds.ac.uk/10.1186/cc2242
Published:
DOI: https://0-doi-org.brum.beds.ac.uk/10.1186/cc2242