- Poster presentation
- Published:
Correlation of continuous cardiac output measured by a pulmonary artery catheter versus impedance cardiography in ventilated patients
Critical Care volume 8, Article number: P58 (2004)
Introduction
Cardiac output can be measured continuously by an invasive pulmonary artery catheter (PAC) or by noninvasive impedance cardiography. The purpose of this study is to determine how well the two methods correlate in ventilated patients.
Method
Patients requiring mechanical ventilation and a PAC were studied in a combined surgical and medical intensive care unit from October 2002 to October 2003. Simultaneous measurements of cardiac output were obtained, utilizing a Swan–Ganz CCOmbo V catheter (Edwards Lifesciences, Irvine, CA, USA) and impedance cardiography utilizing a BioZ (Cardiodynamics, San Diego, CA, USA). Measurements were recorded at initial placement and at 12, 24 36, and 48 hours.
Results
Twenty-two patients were studied. The average PaO2/FiO2 was 184 ± 85. Sixteen (73%) of the patients required dopamine, norepinephrine and/or dobutamine. Overall the correlation coefficient was r = 0.48 and r2 = 0.23. There was a decrease in correlation with time, as shown in Table 1. Six (27%) of the patients had a cardiac output greater than 9 l/min measured by the PAC. The correlation coefficient was r = 0.52 and r2 = 0.27. In the 18 patients with a cardiac output less than 9 l/min, the correlation coefficient was r = 0.89 and r2 = 0.80.
Conclusion
The cardiac output measured by impedance cardiography has good correlation with the PAC in patients with a cardiac output less than 9 l/min. With time, the correlation between the two methods decreases. Further studies are indicated to determine which method is best suited for ventilated patients.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Ziegler, D., Tellez, M., Hess, D. et al. Correlation of continuous cardiac output measured by a pulmonary artery catheter versus impedance cardiography in ventilated patients. Crit Care 8 (Suppl 1), P58 (2004). https://0-doi-org.brum.beds.ac.uk/10.1186/cc2525
Published: