- Perspective
- Open access
- Published:
From pressure to tension: a model of damaging inflation stress
Critical Care volume 27, Article number: 441 (2023)
Abstract
Although the stretch that generates ventilator-induced lung injury (VILI) occurs within the peripheral tissue that encloses the alveolar space, airway pressures and volumes monitor the gas within the interior core of the lung unit, not its cellular enclosure. Measured pressures (plateau pressure, positive end-expiratory pressure, and driving pressure) and tidal volumes paint a highly relevant but incomplete picture of forces that act on the lung tissues themselves. Convenient and clinically useful measures of the airspace, such as pressure and volume, neglect the partitioning of tidal elastic energy into the increments of tension and surface area that constitute actual stress and strain at the alveolar margins. More sharply focused determinants of VILI require estimates of absolute alveolar dimension and morphology and the lung’s unstressed volume at rest. We present a highly simplified but informative mathematical model that translates the radial energy of pressure and volume of the airspace into its surface energy components. In doing so it elaborates conceptual relationships that highlight the forces tending to cause end-tidal hyperinflation of aerated units within the ‘baby lung’ of acute respiratory distress syndrome (ARDS).
Background and objective
Ventilator-induced lung injury (VILI) relates to lung hyperinflation and/or repeated tidal cycles of intolerable mechanical ‘stretch’, which eventually fracture fragile stress-bearing elements or induce lung inflammation [1]. At today’s bedside, driving pressure (DP), plateau pressure (Pplat), positive end-expiratory pressure (PEEP) and tidal volume (VT) are used to monitor that VILI risk. Energy, the area within the envelope that encloses inflation pressure and delivered volume, is required to inflate the lungs but may lead to damage; therefore, ‘power’, the product of frequency and inflation energy per cycle, has drawn increased attention as an integrating indicator of VILI risk [2,3,4]. Mechanical stress is a force-related variable that strains (displaces or deforms) tissue. Although both flow-resistive and non-resistive inflation pressures contribute to dynamic stress and strain, the conserved (so-called ‘elastic’ or static) pressures: Pplat, PEEP and their difference, DP, are the measured pressure components of tidal energy commonly considered to stretch lung tissue [5].
Not all levels of elastic pressure or energy generate excessive power. More specifically, damaging energy recognizes a threshold for monitored pressure (Pt) below which delivered elastic energy can be viewed as ‘safe’ and above which the potential for hazardous overstretch may occur [6]. (Fig. 1) That threshold varies as a function of local transpulmonary pressures and tissue vulnerability to injurious stretch. Yet, even though gas pressure gives rise to them, VILI-inducing mechanical forces are actually generated at the periphery of the shell-like cellular membrane (SLM) that encloses the alveolar space through which gas pressures and volumes are applied and monitored. In concept, these ‘SLM’ stresses and strains relate more directly to changes of surface tension and area than to those of pressure and volume, respectively. (Fig. 2) If so, the tissue-rending force that acts on the cellular SLM and its matrix to initiate VILI may be better reflected by force/unit length, defined as tension than by pressure, which is force/unit area [7]. When considered at the alveolar boundary, neither measured pressures nor tidal volumes directly quantify tidal tissue strains defined as the incremental expansion of surface area. The same might be said of clinically measured energy, power, and each of their component variables measured from within the alveolar airspace.
While the concept of ‘damaging energy’ as determined by crossing of a pressure-defined stress threshold remains highly relevant, the primary purpose of the current work is to refine and complement our previous model of damaging energy [8] to draw a closer connection between the variables routinely measured at the bedside (VT, Pplat, PEEP, DP) and the stresses and strains that actually result from them at the alveolar boundary. Because stretch occurs at the periphery of the alveolar unit and not within the airspace core itself, we propose that, in theory: (1) Tension is the relevant force per unit length that stretches the SLM to strain cellular elements; and (2) The partitioning of elastic energy into tension and surface area helps understand how well measured pressure, volume, and their product approximate ventilatory stress, damaging energy per inflation cycle, and hazardous power [6, 8]. We emphasize that these are mathematical relationships and though relevant, do not intend to mimic the complexity of the true biological state.
Model development
Radial and surface energy
Details that underpin our mathematical model are presented in the online Additional file 1: (Parts 1–4). The first step in developing a conceptually informative yet clinically relatable model from variables routinely measured at the bedside requires important assumptions and simplifications. Prominent among these in our model is to represent the alveolar compartment of the lung considered at any scale as a single expandable, thin-walled and balloon-like hollow sphere of radius (R) with uniform inflation properties. These properties are characterized by a constant compliance (C) applicable to all measured increments of pressure and volume. In actuality, lung units are irregularly shaped and interdependent. Moreover, at low lung volumes even an open unit may unfold as its pressure increases, rather than recoil to store elastic energy [9, 10]. Although such basic geometry is clearly an oversimplification, we note that the shape of the individual alveolus does, in fact, become more spherical as volume increases toward its maximum [9], a process that reduces the ‘corner irregularities’ where stress is focused [9, 11]. While all forms of non-dissipated energy input to the sphere—the conserved or ‘elastic’ energy—must sum to the same value, the partitioning of the components of that total energy can vary independently. [Derivation of this equivalence is provided in the Additional file 1: Part 1]. Using a highly simplified spherical model allows translation of total radial elastic energy, (a measurable pressure (P)—volume (V) quantity), into surface energy, (a circumferential tension (T)—surface area (As) quantity). This radial to surface energy conversion permits estimation of the relevant stresses and strains within the cellular SLM.
The incremental elastic energy needed for inflation during tidal inflation is stored under tension within the expanding margin of the hollow sphere where the cellular elements reside; conceptually, therefore, increments of circumferential tension (T) and surface area (As) determine the tissue stretch that stimulates VILI. In such a basic model, the stressing variable relevant to the alveolar tissue is tension (T), not simply the directly accessible intracavity pressure measured within the alveolus and ventilator circuitry. Technically, T is not formally defined as stress because T is a force/length, whereas stress is force/area [7]. Tension at the surface is generated by forces developed tangential to the radius (R) of the sphere. In accordance with the ‘law of LaPlace’ for such thin-walled expandable spheres: T = PR/2, where P is the sphere’s internally measured pressure [12] (Fig. 2). Consequently, the greater the diameter of an individual spherical lung unit, the greater its SLM tension for a given pressure. This relationship applies to spheres of any dimension—whether large or small. Importantly, unlike the equivalence of input elastic energy to increments of surface elastic energy, the subcomponents of these identical energies differ. Yet, under static, no-flow conditions, radial forces resulting from P must be exactly counterbalanced by the net inward vector of non-radial (tangentially orthogonal) surface forces that arise from circumferential tension–what are often known as ‘hoop stresses’ [13] (Fig. 2). This mandate to balance the outward force of a given pressure with the inwardly directed force generated by circumferential tension holds true whatever the sphere’s radius. The smaller the sphere’s radius and sharper its surface curvature, the greater the leverage of tension to generate that inward force vector (Fig. 2). Reduced leverage angle for generating its inward force vector requires tension to rise disproportionately to area as the alveolus distends.
With the difference between T and P as well as the equivalence of surface energy and radial energy in mind, it follows that strain imposed by inflation on the cellular SLM is not the simple relationship of volume increment (∆V) to resting volume (V = 4πR3/3), but a more direct correlate of the incremental change in the surface area of the SLM (As). This SLM area is As = 4πR2. Because the relationship of T and AS quantifies the incremental elastic energy stored in the SLM during inflation, alternative estimates of energy input integrating P and V or T and As are identical, but in these estimates, ∆V ≠ ∆As. (Derivation of their quantitative relationship is provided in the Additional file 1: Parts 1 and 2.) As already noted, assuming perfect efficiency of the translation of non-dissipated airspace energy to surface elastic energy, ∆T must differ from ∆P during expansion to maintain equivalence between input radial and SLM-stored elastic energies. How they differ relates to R and its incremental change: \(\Delta T=\frac{1}{2}R\Delta P+\frac{1}{2}P\Delta R\).
Role of unstressed volume (FRC) in components of surface elastic energy
For a sphere, the two key variables defining T are P and R, the first of which can be directly measured in clinical practice and the other only estimated. (The Additional file 1: part 3 outlines the rationale underpinning that estimation.) In broad outline, the value of R for our lung model derives from its absolute volume expressed in clinical terms: V = FRC + PEEPxC + VT. With the resting volume, or functional residual capacity, FRC, estimated as described in Additional file 1: part 3, R of a sphere at either the macro (lung) or micro (subunit) scales is determined from its corresponding absolute volume VA as: R = (3VA/4π)1/3. This R can then be used to estimate the corresponding ‘SLM’ tensions (T = PR/2) and ‘SLM’ areas, As = 4πR2 of interest.
In a clinical context, VA at end-expiration equals its unstressed FRC plus any PEEP-related volume increment (VPEEP). The VA at end-inspiration is the sum of FRC, VPEEP and VT. In practice, the FRC of an ARDS ‘baby lung’ typically goes unmeasured but bears an inverse relationship to the degree of ARDS severity [14]. Whatever the severity, the baby lung of ARDS is assumed by convention to be comprised of a number (n) of normally compliant individual alveolar subunits [14, 15]. Under this assumption, measured tidal compliance (Cobs) is a direct function of the number of quasi-normal subunits, not of subunit stiffness. We note here, however, that while their expansion properties may be similar, the vulnerability of a heathy lung unit to stretching injury is likely much less than a pre-injured unit. Moreover, immersed in a heterogeneous mechanical environment, the latter are also exposed to the amplified forces of stress focusing. It follows that the unmeasured FRC is highly VILI-relevant, as it correlates inversely with individual subunit diameter at end-inflation and the resulting stretch that occurs as the n units of the baby lung accommodate and share the tidal volume (VT). Though unmeasured in the individual patient, a reasonable (though imprecise) estimate of the baby lung’s resting and unstressed volume without any PEEP applied (Vrest) might be: Vrest = (Cobs/Cnorm) \(\times\) FRCpred, where predicted normal FRC (FRCpred) is a known function of age, gender, and lean body mass during health [16, 17]. The Cobs of the respiratory system is directly calculated from monitored tidal pressures and tidal volume. Normal compliance (Cnorm) varies modestly with body position but generally ranges from 80–100 ml/cmH2O [17].
Comparison within and between baby lungs of different capacities
The LaPlace equation dictates that expansion of any given sphere increases its wall tension. For a spherical structure of volume (V), however, the tension (T) that develops within its spherical subunits will depend upon the number (n) of such units that comprise their collective volume (V) (Additional file 1: Part 2). We note that when expressed in the same numerical units, a larger sphere that contains the same net volume as a collection of smaller subunits will have a SLM tension that exceeds the wall tensions of its constituent subunits (Fig. 3). More specifically, a simple calculation indicates that for two non-communicating collections (A and B) of identical cumulative volume containing nA and nB numbers of subunits, respectively:
(Additional file 1: 2)In other words, the same absolute volume contained in a separate sphere comprised of fewer units will have higher SLM subunit tension; therefore, increasing the capacity (subunit numbers) of the hypothetical baby lung will reduce tensions in the individual subunits that comprise it.
Accounting for wall thickness and subunit stiffness
To this point, the SLM of the sphere has been considered of negligible thickness. But in fact, alveolar walls do have thickness comparable to their radii. Discounting any change of capillary blood volume, a biological alveolar ‘sphere’ would have a fixed quantity of tissue that thins with stretching. In this setting, the modified LaPlace equation applies: T = PR/2z, where z represents wall thickness. Starting from a SLM with an unchanging amount of cellular tissue, wall thickness will increase as the sphere contracts and decrease with its expansion. (Fig. 4) Therefore, at high volumes, T is increased not only by higher P and R, but to some extent by lower z values. Along this line, within the same lung, modeled alveoli of smaller radii (e.g., gravitationally dependent) should have lower wall tensions than non-dependent ones at the same pressure, due at least in part to their greater wall thickness. For the same thickness and internal volume, a stiffer alveolar unit or one amidst non-inflating tissues that buttress it (e.g., in late-stage ARDS) would experience less tension and stretch less in response to a pressure increase. For these, the pressures that demarcate the threshold for damaging energy would theoretically increase.
Model predictions
Allowing for its basic assumptions, this simple model yields predictions of potential interest to questions that surround VILI avoidance. These include:
-
1.
Within a given baby lung, SLM tension of its constituent lung units increases disproportionately to measured pressure, assuming C remains unchanged.
-
2.
For set values of VT and PEEP, end-inspiratory SLM tension (PR/2), the rending (or ‘ripping’) force per unit length, relates inversely to FRC at zero PEEP.
-
3.
Strain resulting from non-dissipated (‘elastic’) energy during inflation from R1 to R2 may differ from that estimated from measured tidal volumes and pressures because ∆ surface area is not identical to ∆ volume (Additional file 1: Part 4):
$$\frac{\Delta A}{\Delta V}=[3\left({{R}_{2}}^{2}-{{R}_{1}}^{2}\right)]/({{R}_{2}}^{3}-{{R}_{1}}^{3})$$Consequently, tension rises in direct relation to baby lung unit volume, but hazardous energy causing excessive strain (stretch) relates more closely to relative changes in surface area than to those in volume. What’s more, as volume increases, the input energy causes T (the stressing factor of SLM energy) to rise faster than AS (the strain factor), thereby altering the pressure threshold for damaging energy.
-
4.
Lung unit stiffness, SLM thickness, or smaller unit dimensions lessen the tension developed in response to a given pressure increment. Simultaneously, the discrepancy narrows between the variables indicating stress (pressure vs. tension).
-
5.
As the cellular SLM thins during expansion, increments of damaging energy accentuate the tendencies for alveolar coalescence and cystic transformation, because ∆AS and ∆T both rise for a given ∆P. Thinning of the SLM accentuates this tendency.
Clinical relevance
This highly simplified modeling exercise is not intended for direct and quantitative clinical translation. In opening a discussion of tension based on single isolated spherical structures, we understand that the architecture and micromechanics of the injured lung are far more complex. Nonetheless, consideration of the shared fundamental physical principles of the model may help instruct bedside respiratory support. At the alveolar periphery, the damaging portion of ∆ PV energy translates into increments of tension and area, which logically are the actual generators and correlates of hazardous stress and strain. Tidal increments of energy weigh progressively toward its SLM tension component. We note here that rising tension is not always detrimental; until a damage threshold is crossed [6, 8], higher tension may be well tolerated or help in the recruitment of unstable units [11]. Once beyond that threshold, however, unnecessary volume increments are to be avoided. Minimizing disparities of regional volumes and tensions, as by prone positioning, is also advisable.
Awareness of lung unit hyperinflation during ARDS has received increased recent attention [18,19,20]. Unmeasured end-expiratory volume per aerated lung unit is a key determinant of cellular stretch during VT delivery. In that context, a sharper focus on disproportionately rising SLM tension underscores the importance of avoiding end-tidal overdistention by lessening mechanical heterogeneity and avoidance of excessive tidal volumes or unwarranted high PEEP. Even when recommended guidelines are followed, end-tidal overstretching of the small ‘baby lung’ may help account for the ‘paradoxical responses’ to chest wall compression and for more frequent barotrauma in ARDS-associated Covid-19 pneumonia [21]. Localized overexpansion increases vascular resistance, redirects blood flow, and disrupts optimal ventilation/perfusion matching. More VILI-relevant, escalating tension coupled with SLM thinning during expansion increases the risk of disruption and of coalescence of multiple small units into fewer but larger ones containing the same collective volume, raising tension at the airspace periphery (Fig. 3). This tendency to merge alveolar subunits accentuates when the stress-bearing extracellular matrix elements and microstructures have been degraded by inflammation, encouraging rapid formation of what has been termed ‘tension cysts’ [22]. Rarely, these high airspace and vascular stresses, acting together, have the potential to disrupt the gas to blood barrier and generate systemic gas emboli. [23].
While extensive subunit overdistention must be avoided, our model also underlines why recruitment at safe plateau pressures is desirable, as adding functional units not only reduces driving pressure but also reduces the end-tidal alveolar tension developed during breath delivery. Conversely, advancing disease or VILI reduce the number of aerated units, shrinking the ‘baby lung’, thereby raising subunit surface energy, encouraging end-tidal hyperinflation, and predisposing to cystic transformation [21, 24, 25].
Model limitations
While perhaps conceptually valid in its basic premise, such a simplified mathematical model clearly has limited correspondence with the complex geometry that characterizes the actual biological environment of the injured lung. In our model’s unmodified form, La Place’s ‘law’ for thin-walled spheres must apply to all compartment sizes, and wall thickness is not considered. In vivo, however, neither the integrated lung and nor its constituent units are negligibly thin, mechanically independent identical spheres exposed to a single transpulmonary pressure. In actuality, the latter vary locally in accordance with their immediate local environments, causing regional differences of resting subunit volumes [10]. Moreover, alveolar walls are relatively thick in relation to their radii, and biological lung unit contours are irregular and interdependent, with highly variable topography of stress-focused corners and interfaces [9, 10]. Importantly, the assumption of quasi-normal specific compliance of all aerated baby lung subunits, while perhaps initially reasonable, is open to serious question in the later stages of advancing ARDS, when organizing inflammation and fibrotic processes are underway. In every phase of ARDS, constancy of the pressure–volume relationship (unchanging subunit compliance) seems unlikely to apply at the extremes of the inflation range. Finally, the perception itself that individual alveoli alter their radii after opening remains somewhat controversial [10], as does the structure that actually expands—alveolar ducts or sacs (10). While unfolding that recruits SLM may occur at low dimensions, once unfolded, tension rises as a direct function of pressure at greater dimensions, whatever the shape.
Summary and conclusion
The stretch that generates VILI occurs in the cellular enclosure (SLM) of the alveolar space, while airway pressures and volumes monitor the interior core. Though clinically useful, Pplat, DP and VT paint an incomplete picture of SLM forces and neglect the partitioning of applied tidal energy into the increments of tension and change of surface area occurring at the periphery. The absolute lung unit volume, influenced by resting FRC as well as by PEEP and VT, influences SLM tension. The resulting forces applied to the SLM are more tangential than radial, and these directly influence the actual stresses and strains experienced by the alveolar cells at risk for VILI, especially at points of discordant compliance–stress-focused interfaces. Because the radius of the lung subunit is a fundamental component of tension, reduction in their number promotes the tendency for end-tidal hyperinflation and eventual microcyst formation. For these aspects of VILI, the ‘elastic’ pressure and tidal volume products that currently define damaging energy (DE) and power (f x DE) need further refinement to consider the tension x area product that comprises elastic energy applied and stored at the alveolar boundary.
All authors reviewed and agreed with the final version of this manuscript.
Abbreviations
- ARDS:
-
Acute respiratory distress syndrome
- A S :
-
Surface area
- C :
-
Compliance
- C norm :
-
Predicted compliance
- C obs :
-
Observed compliance
- DE:
-
Damaging energy
- F :
-
Force
- DP:
-
Driving pressure (Pplat minus PEEP)
- FRC:
-
Functional residual capacity (unstressed resting volume)
- N :
-
Number
- P :
-
Measured pressure
- PEEP:
-
Positive end-expiratory pressure
- P plat :
-
Plateau pressure (end-inspiratory elastic pressure)
- P t :
-
Threshold pressure for damage
- R :
-
Sphere radius
- T :
-
Tension at the sphere’s surface
- SLM:
-
Shell-like membrane enclosing the alveolar sphere
- V :
-
Volume
- V A :
-
Absolute (total) volume
- VILI:
-
Ventilation-induced lung injury
- V Peep :
-
Volume increment due to PEEP
- V Rest :
-
Volume at rest (at FRC)
- V T :
-
Tidal volume
References
Slutsky AS, Ranieri VM. Ventilator-induced lung injury. N Engl J Med. 2013;369(22):2126–36. https://0-doi-org.brum.beds.ac.uk/10.1056/NEJMra120870.
Cressoni M, Gotti M, Chiurazzi C, Massari D, Algieri I, Amini M, Cammaroto A, Brioni M, Montaruli C, Nikolla K, Guanziroli M, Dondossola D, Gatti S, Valerio V, Vergani GL, Pugni P, Cadringher P, Gagliano N, Gattinoni L. Mechanical power and development of ventilator-induced lung injury. Anesthesiology. 2016;124:1100–8.
Santos RS, Maia L, Oliveira MV, Santos CL, Moraes L, Pinto EF, Samary CS, Machado JA, Carvalho AC, Fernandes MV, Martins V, Capelozzi VL, Morales MM, Koch T, de Gama Abreu M, Pelosi P, Silva PL, Rocco PR. Biological impact of mechanical power at high and low tidal volumes in experimental mild acute respiratory distress syndrome. Anesthesiology. 2018;128(6):1193–206. https://0-doi-org.brum.beds.ac.uk/10.1097/ALN.0000000000002143.
Serpa Neto A, Deliberato RO, Johnson AEW, Bos LD, Amorim P, Pereira SM, Cazati DC, Cordioli RL, Correa TD, Pollard TJ, Schettino GPP, Timenetsky KT, Celi LA, Pelosi P, Gama de Abreu M, Schultz MJ. Mechanical power of ventilation is associated with mortality in critically ill patients: an analysis of patients in two observational cohorts. Intensive Care Med. 2018;44:1914–22.
Fan E, Del Sorbo L, Goligher EC, Hodgson CL, Munshi L, et al. An Official American Thoracic Society/European Society of Intensive Care Medicine/Society of Critical Care Medicine Clinical Practice Guideline: Mechanical Ventilation in Adult Patients with Acute Respiratory Distress Syndrome. Am J Respir Crit Care Med. 2017;195(9):1253–63. https://0-doi-org.brum.beds.ac.uk/10.1164/rccm.201703-0548ST.
Marini JJ, Rocco PRM. Which component of mechanical power is most important in causing VILI? Crit Care. 2020;24:39. https://0-doi-org.brum.beds.ac.uk/10.1186/s13054-020-2747-4.
Merriam-Webster's Collegiate Dictionary, 11th ed., Merriam-Webster, 2003, Springfield Massachusetts, (USA)
Marini JJ, Thornton LT, Rocco PRM, Gattinoni L, Crooke PS. Practical assessment of risk of VILI from ventilating power: a conceptual model. Crit Care. 2023;27:157. https://0-doi-org.brum.beds.ac.uk/10.1186/s13054-023-04406-9.
Roan E, Waters CM. What do we know about mechanical strain in lung alveoli? Am J Physiol Lung Cell Mol Physiol. 2011;301:L625–35.
Knudsen L, Ochs M. The micromechanics of lung alveoli: structure and function of surfactant and tissue components. Histochem Cell Biol. 2018;150:661–76. https://0-doi-org.brum.beds.ac.uk/10.1007/s00418-018-1747-9.
Mead J, Takishima T, Leith D. Stress distribution in lungs: a model of pulmonary elasticity. J Appl Physiol. 1970;28:596–608.
Shelmerdine SC, Barber JL, George CD. Applications of Laplace’s law in clinical medicine: a radiological pictorial review. Br J Hosp Med (Lond). 2013;74(8):451–6. https://0-doi-org.brum.beds.ac.uk/10.12968/hmed.2013.74.8.451.
Ibrahim A, Ryu Y, Saidpour M. Stress analysis of thin-walled pressure vessels. Mod Mech Eng. 2015;5:1–9.
Gattinoni L, Pesenti A. The concept of “baby lung.” Intensive Care Med. 2005;31(6):776–84. https://0-doi-org.brum.beds.ac.uk/10.1007/s00134-005-2627-z.
Gattinoni L, Marini JJ, Pesenti A, Quintel M, Mancebo J, Brochard L. The, “baby lung” became an adult. Intensive Care Med. 2016;42(5):663–73. https://0-doi-org.brum.beds.ac.uk/10.1007/s00134-015-4200-8.
Withers RT, Lemmey AB. Lung volume and spirometric standards for healthy female lifetime nonsmokers. Hum Biol. 1989;61(3):343–68.
Lumb AB. Nunn’s applied respiratory physiology. 8th ed. Edinburgh: Elsevier; 2017. p. 51–3.
Terragni PP, Rosboch G, Tealdi A, et al. Tidal hyperinflation during low tidal volume ventilation in acute respiratory distress syndrome. Am J Respir Crit Care Med. 2007;175(2):160–6.
Rezoagli E, Bastia L, Brochard L, et al. Physical manoeuvres in patients with ARDS and low compliance: bedside approaches to detect lung hyperinflation and optimise mechanical ventilation. Eur Respir J. 2023;61:2202169. https://0-doi-org.brum.beds.ac.uk/10.1183/13993003.02169-2022.
Selickman J, Marini JJ. Chest wall loading in the ICU: pushes, weights, and positions. Ann Intensive Care. 2022;12(1):103. https://0-doi-org.brum.beds.ac.uk/10.1186/s13613-022-01076-8.
Steinberger S, Finkelstein M, Pagano A, Manna S, Toussie D, Chung M, et al. Barotrauma in COVID 19: incidence, pathophysiology, and effect on prognosis. Clin Imaging. 2022;90:71–7. https://0-doi-org.brum.beds.ac.uk/10.1016/j.clinimag.2022.06.014.
Albelda SM, Gefter WB, Kelley MA, et al. Ventilator-induced subpleural air cysts: clinical, radiographic, and pathologic significance. Am Rev Respir Dis. 1983;127:360–5.
Marini JJ, Culver BH. Systemic gas embolism complicating mechanical ventilation in the adult respiratory distress syndrome. Ann Intern Med. 1989;110(9):699–703. https://0-doi-org.brum.beds.ac.uk/10.7326/0003-4819-110-9-69.
Marini JJ, Gattinoni L. Time course of evolving ventilator-induced lung injury: the “shrinking baby lung.” Crit Care Med. 2020;48:1203–9.
Kalenka A, Gruner F, Weiß C, Viergutz T. End-expiratory lung volume in patients with acute respiratory distress syndrome: a time course analysis. Lung. 2016;19:527–34. https://0-doi-org.brum.beds.ac.uk/10.1007/s00408-016-9892-1.
Funding
PRMR is supported by Brazilian Council for Scientific and Technological Development (CNPq) and Carlos Chagas Filho, Rio de Janeiro State Foundation (FAPERJ).
Author information
Authors and Affiliations
Contributions
JJM conceived the project, collaborated in manuscript development, and supervised the varied phases of this work. LTT was closely involved with the development of the project and manuscript at all stages. PRMR provided valued intellectual input and aided in manuscript development. PC worked closely with JJM in developing the core concepts, provided vital intellectual input and assured the accuracy of its predictive equations. All authors reviewed and agreed with the final version of this manuscript.
Corresponding author
Ethics declarations
Ethics approval and consent to participate
Not applicable.
Consent for publication
Not applicable.
Availability of data and materials
Not applicable.
Competing interests
The authors declare no competing interests.
Additional information
Publisher's Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Supplementary Information
Additional file 1
. Rationale and detailing of the mathematical model.
Rights and permissions
Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Marini, J.J., Thornton, L.T., Rocco, P.R.M. et al. From pressure to tension: a model of damaging inflation stress. Crit Care 27, 441 (2023). https://0-doi-org.brum.beds.ac.uk/10.1186/s13054-023-04675-4
Received:
Accepted:
Published:
DOI: https://0-doi-org.brum.beds.ac.uk/10.1186/s13054-023-04675-4