Feasibility of 3D-EIT in identifying lung perfusion defect and V/Q mismatch in a patient with VA-ECMO

To the Editor,

Three-dimensional (3D) electrical impedance tomography (EIT) is a recently developed lung ventilation and perfusion imaging technique. Conventional two-dimensional (2D) EIT provides information on a single plane, which potentially omits important information concerning changes in global lung conditions. With the advantage of providing lung information on three planes, 3D-EIT may overcome these pitfalls. Recently, the saline bolus method of 3D-EIT has been used to assess regional lung perfusion and ventilation/perfusion (V/Q) mismatch in an animal study [1]. In patients under extracorporeal membrane oxygenation (ECMO) therapy, bedside assessment of lung perfusion remains a major challenge for either 2D or 3D-EIT techniques.

In this study, we demonstrate the feasibility of a novel 3D-EIT method for lung perfusion assessment and its first clinical application in a post-cardiac surgery patient with veno-arterial extracorporeal membrane oxygenation (VA-ECMO) treatment.

Case presentation

A 63-year-old female patient was admitted with multiple space-occupying tumors in the main and right pulmonary artery, which were subsequently confirmed to be intimal sarcoma. During tumor resection surgery, the patient presented with significantly decreased PaO2/FiO2 ratio and blood pressure and was difficult to wean off cardiopulmonary bypass. Hence, she received peripheral VA-ECMO treatment and was transferred to the ICU for further management. The speed of ECMO was 3500 rpm with blood flow of 3–3.5 L/min. Radiological imaging results are presented in Fig. 1.

Fig. 1
figure 1

Transesophageal echocardiography (TEE), computed tomography (CT), and saline bolus-based 3D electrical impedance tomography (EIT) results. The EIT measurement system adopted alternating currents with a rms of 2 mA and a frequency of 20 kHz. Voltages were measured at a rate of 20 frames per second. Details of the injection-measurement pattern can be found in [2] (IV-A). The image reconstruction algorithm was run on one NVIDIA Tesla V100 GPU card and the image was produced at a rate of 0.125 frames per second. A TEE image of the mid-esophageal bicaval view. The red arrow indicates a massive embolism in the right atrium. B TEE image of the mid-esophageal ascending aortic short-axis view. The red arrow demonstrates that the color Doppler found no blood perfusion signal in the right pulmonary artery. C Chest CT image indicated that the lung was almost normal. D Lung ventilation image of EIT in coronary, sagittal, and axial planes with low-ventilated regions marked in dark blue and high-ventilated regions in light grey. E Lung perfusion image of EIT in coronary, sagittal, and axial planes. Regions with low perfusion are displayed in dark red and high perfusion in white. F Ventilation/perfusion (V/Q) image of EIT in coronary, sagittal, and axial planes demonstrating the regional V/Q distribution. Regions of dead space (high ventilation and low perfusion) are shown in dark gray. Shunt regions (low ventilation and high perfusion) are marked in red. Areas of matched V/Q ratios are presented in yellow. The cardiac output (2.5 L/min) and minute volume (6 L) were included in the V/Q calculation according to François Perier’s study [3]. G The percentage of deadspace, shunt, and V/Q match pixels in 3D global and different planes. LA Left atrium, RA Right atrium, RPA Right pulmonary artery, AO Aorta, MPA Main pulmonary artery, Axi Axial plane, Cor Coronary plane, Right Sag Right sagittal plane, Left Sag Left sagittal plane

EIT methods

After obtaining informed written consent from the next of kin, 3D-EIT was performed on ICU day 34 for lung assessment. Two belts with 16 surface electrodes each were placed at the level of the 3rd and 5th intercostal spaces. Data were recorded by an EIT device (Infivision 1900, CHINA). Lung perfusion was evaluated by saline bolus-based and pulsatility-based EIT methods. A 20 ml 10% saline bolus was rapidly injected via a central venous catheter during a breathing pause [4]. Images of the pulsatility method were generated based on a separated cardiac-related signal [2].

EIT results

A significant V/Q mismatch was detected on the right lung with a deadspace of 71.03% in the right sagittal plane and 41.81% in 3D global (Fig. 1D–G). 3D-EIT revealed that ventilation was almost normal across the lung (Fig. 1D). However, regarding perfusion, a severe defect was observed in the right lung (Fig. 1E). With conventional 2D-EIT, abnormal perfusion could also be identified. Dynamic 3D V/Q videos are present in Additional file 1: Video S1, Additional file 2: Video S2 and Additional file 3: Video S3. We extracted data on regional ventilation, perfusion, and V/Q ratios from the coronary, sagittal, and axial planes (Fig. 1D–G).

Comparison of perfusion images using saline bolus and pulsatility methods is shown in Additional file 4: Fig. S1. In the pulsatility strategy, images varied and were more dependent on the slices selected. Whereas the saline bolus-based method provided consistent results of perfusion defect across different planes and slices.

Discussion

The classical single-belt 2D-EIT has certain pitfalls. It lacks spatial information. Considering the significant inhomogeneity of V/Q distribution in patients under pathologic pulmonary conditions, 2D-EIT fails to capture crucial V/Q information across the lung at coronary and sagittal levels. With more information available, 3D-EIT could potentially provide a more accurate profile for individualized lung assessment. Furthermore, in the single-plane EIT, variations in the sensitive region across the thorax and out-of-plane changes in the conductivity of the volume center could affect the results [1]. Moreover, changes in electrode belt position could significantly influence the result in 2D-EIT, which increases uncertainty when analyzing data between patients. Whereas the double-belt design in 3D-EIT can reduce the impact.

Both saline bolus and pulsatility-based approaches have been adopted to investigate lung perfusion in 2D-EIT. A previous experimental study demonstrated that the saline-based strategy was superior to the pulsatility strategy [5], which is consistent with the present study. For real-time and accurate monitoring of 3D lung perfusion, combining both strategies might be the way forward.

It remains a great challenge for bedside global lung perfusion assessment during VA-ECMO treatment. The flow of VA-ECMO could impact the assessment by directly transferring the injected saline from the femoral vein to the artery [4]. To minimize this effect, we used 20 ml 10% NaCl and lowered ECMO flow to 2L/min during the contrast procedure. A significantly decreased impedance caused by saline injection was observed and the 3D-EIT result was consistent with the clinical judgment and result of TEE.

Yet before the era of 3D-EIT may arrive, certain challenges must be overcome. Firstly, the double-belt design increases inconvenience in clinical procedures. Secondly, the 3D algorithm is complex and time-consuming. With the improvement of hardware and algorithms, real-time bedside 3D-EIT monitoring could be achieved in the near future. Thirdly, further evidence is necessary for continued validation of the efficacy of 3D-EIT and its advantages over 2D-EIT in clinical practice.

In conclusion, 3D-EIT is able to detect lung perfusion defects in a patient with ECMO treatment. It might be a promising new technique for bedside monitoring of V/Q changes in different dimensions.

Availability of data and materials

The datasets used and/or analyzed during the current study are available from the corresponding author on reasonable request. The datasets supporting the conclusions of this article are included within the article and its additional files.

Abbreviations

EIT:

Electrical impedance tomography

2D:

Two-dimensional

3D:

Three-dimensional

V/Q:

Ventilation/perfusion

VA-ECMO:

Veno-arterial extracorporeal membrane oxygenation

TEE:

Transesophageal echocardiography

CT:

Computed tomography

References

  1. Larrabee S, Nugen S, Bruhn A, et al. Three-dimensional electrical impedance tomography to study regional ventilation/perfusion ratios in anesthetized pigs. Am J Physiol Lung Cell Mol Physiol. 2023;325(5):L638–46.

    Article  CAS  PubMed  Google Scholar 

  2. Zhang K, Wang L, Guo R, Lin Z, Li M, Yang F, Xu S, Abubakar A. A deep generative model-integrated framework for 3-D time-difference electrical impedance tomography. IEEE Trans Instrum Meas. 2022;72:1–12.

    CAS  Google Scholar 

  3. Perier F, Tuffet S, Maraffi T, et al. Effect of positive end-expiratory pressure and proning on ventilation and perfusion in COVID-19 acute respiratory distress syndrome. Am J Respir Crit Care Med. 2020;202(12):1713–7.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  4. He H, Jiang J, Xu M, et al. Saline bolus-based electrical impedance tomography method for rapid bedside assessment of regional lung perfusion during ECMO therapy. Crit Care. 2022;26(1):266.

    Article  PubMed  PubMed Central  Google Scholar 

  5. Borges JB, Suarez-Sipmann F, Bohm SH, et al. Regional lung perfusion estimated by electrical impedance tomography in a piglet model of lung collapse. J Appl Physiol (1985). 2012;112(1):225–36.

    Article  PubMed  Google Scholar 

Download references

Acknowledgements

Not applicable.

Funding

This work was supported by the National High-Level Hospital Clinical Research Funding (2022-PUMCH-D-005), the National Natural Science Foundation of China (Grant No 82272249), and the National Key Research and Development Program (2022YFC2404805).

Author information

Authors and Affiliations

Authors

Contributions

HH, YG, KZ, ML, SY, QW, YC, YL, and ZZ conceived and designed the protocol. HH, YG, SY, QW, YC, and YL collected study data. KZ and ML participated in data interpretation. YG and HH drafted the manuscript. HH, YG, KZ, ML, SY, QW, YC, YL and ZZ revised the manuscript. All authors have given final approval of the submitted and published versions.

Corresponding author

Correspondence to Huaiwu He.

Ethics declarations

Ethics approval and consent to participate

Informed written consent was obtained from the patient’s next of kin before the examination.

Consent for publication

Consent for publication has been obtained from the patient’s next of kin.

Competing interests

The authors declare that they have no competing interest.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Supplementary Information

Additional file 1: Video S1. 3D-EIT video in lung perfusion.

Additional file 2: Video S2. 3D-EIT video in lung ventilation.

Additional file 3: Video S3. 3D-EIT video in Ventilation/perfusion (V/Q) match.

Additional file 4: Figure S1.

The comparison of saline bolus-based and pulsatility-based methods for accessing lung perfusion. Regions with low perfusion are marked in dark red and high perfusion in white. In the axial plane, lung perfusion images were generated from top to bottom (slices 1 to 20) with an interval of 12mm between each slice. Slice 1 and slice 20 were 23cm apart. The upper and lower belts were at the level of 6 to 7 and 14 to 15 slices respectively. A Saline bolus-based electrical impedance tomography (EIT) method in the axial plane. In slices 1 to 2 and 18 to 20, the perfusion signal was vague. B Saline bolus-based EIT lung perfusion images in coronary, sagittal, and axial planes. C Pulsatility-based EIT lung perfusion images from top to bottom in the axial plane. D Pulsatility-based EIT lung perfusion images in coronary, sagittal, and axial planes. The perfusion signal was able to be detected from slices 1 to 18. In slices close to the lung apices and lung base, abnormal perfusion of the right lung could be noted. Yet in other slices, absent perfusion in the right lung cannot be observed. Axi Axial plane, Cor Coronary plane, Right Sag Right sagittal plane, Left Sag Left sagittal plane.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Gao, Y., Zhang, K., Li, M. et al. Feasibility of 3D-EIT in identifying lung perfusion defect and V/Q mismatch in a patient with VA-ECMO. Crit Care 28, 90 (2024). https://0-doi-org.brum.beds.ac.uk/10.1186/s13054-024-04865-8

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://0-doi-org.brum.beds.ac.uk/10.1186/s13054-024-04865-8