Effect of a self-developed fixation device on preventing endotracheal intubation-related pressure injury: a randomised controlled trial

Abstract

Objective

To evaluate the effects of our self-developed endotracheal tube fixation device in mechanically ventilated patients.

Methods

In a dual-centre randomised controlled trial, patients who were expected to require mechanical ventilation for over 48 h were assigned to the observation group (using self-developed device) or the control group (using the traditional device). The primary endpoint was the incidence of endotracheal intubation-related pressure injury (EIRPI).

Results

Fifty-one patients in the observation group and 54 patients in the control group were analysed. The incidence of EIRPI was 7.8% in the observation group and 33.3% in the control group (p = 0.001). Lip pressure injury (PI) occurred in 0 versus 14 (25.9%) patients in the observation versus control groups (p < 0.001). Both oral–mucosal and facial PIs were similar between the two groups.

Conclusions

The use of the novel device reduced the incidence of EIRPI, especially lip PI.

Trial registration Chinese Clinical Trial Registry ChiCTR2300078132. Registered on 29 November 2023

Introduction

Endotracheal intubation through the oral cavity is the main method used to establish an artificial airway in critically ill patients [1, 2]. Effective fixation of the endotracheal tube is crucial to ensure respiratory function. Common clinical methods for fixing endotracheal tubes include tape, ropes, and commercially available fixation devices, which often lead to complications such as lip pressure injury (PI) [3] and unplanned extubation [4]. These complications cause suffering, prolong hospital stays, and increase healthcare expenses [5,6,7]. Therefore, improving the design of fixation devices is of considerable importance.

Thus, our team developed a novel device, which was authorised by the China National Intellectual Property Administration (patent no.: ZL202221745523.5). In this prospective study, we aimed to evaluate the effects of the novel device in mechanically ventilated patients. To the best of our knowledge, randomised controlled trials comparing PIs associated with endotracheal tube fixation devices from a design perspective are rare.

Methods

Study design

This prospective dual-centre randomised controlled trial aimed to compare the effect of a novel device (Fig. 1A) versus a traditional device (Zhejiang Haisheng Medical Equipment Co., Ltd.) (Fig. 1B) among critically ill adults requiring intubation and mechanical ventilation for at least 48 h. This study was approved by the Institutional Review Board of Nanjing Lishui People’s Hospital and the Medical Ethics Committee of Xuzhou City Central Hospital.

Fig. 1
figure 1

A Structure diagram of the self-developed endotracheal tube fixation device. B Structure diagram of the traditional endotracheal tube fixation device. C Photograph of the patient wearing the self-developed endotracheal tube fixation device. D Photograph of the patient wearing the traditional endotracheal tube fixation device

Study participants

Patients admitted to the intensive care unit (ICU) of the two hospitals from 14 November 2022 to 20 October 2023 who required endotracheal intubation and met the study’s inclusion criteria were randomly allocated to two groups: the observation group with the novel device (Fig. 1C) and the control group with the traditional device (Fig. 1D).

The inclusion criteria were as follows: (1) age ≥ 18 years; (2) meeting the indications for endotracheal intubation; and (3) signing the informed consent by themselves or their family member.

The exclusion criteria were as follows: (1) patients who were not suitable for orotracheal intubation; (2) patients with damaged lip, facial skin, or oral mucosa before intubation; and (3) patients who were not suitable for wearing headbands.

Withdrawal criteria were as follows: (1) patients with intubation time/mechanical ventilation time < 48 h, (2) patients who abandoned the treatment during the trial, and (3) patients or family members who voluntarily requested withdrawal during the trial.

Randomisation

Patients were randomly allocated to the observation or control group by a person other than the researchers in a 1:1 ratio with random sequences generated using IBM SPSS Statistics for Windows, version 25.0 (IBM Corp., Armonk, NY, USA), stratified by centre. Allocation information was concealed using sealed opaque envelopes.

Study outcomes

The primary outcome was the appearance of any endotracheal intubation-related pressure injury (EIRPI) that occurred on the lips, oral mucosa, or facial skin during intubation. Secondary outcomes included duration of mechanical ventilation, tube dislodgement, and ventilator-associated pneumonia (VAP).

Statistical analyses

We estimated the expected incidence of EIRPI to be 28% in the control group and 5% in the observation group based on the data from our preliminary test. The sample size was calculated using PASS software, version 15.0.5 (NCSS, LLC., Kaysville, Utah, USA) with 90% power and a two-sided alpha level of 0.05. Considering a 10% dropout rate, we planned to randomise 112 patients to achieve the 100 eligible patients required for our study.

All hypothesis tests were two-sided, and the threshold for significance was set to 0.05. The Shapiro–Wilk test was used to assess the normality of the distribution. Non-normally distributed continuous variables were presented as medians (quartile 1 and quartile 3), and the Mann–Whitney U-test was used to assess differences between the two groups. The proportions of categorical variables were compared using the Chi-squared test or Fisher’s exact test, as appropriate. All analyses were performed using IBM SPSS Statistics for Windows, version 25.0.

Results

Baseline characteristics

Inclusion, exclusion, and withdrawal criteria resulted in 51 patients in the observation group and 54 patients in the control group (Fig. 2). Baseline characteristics between the two groups showed no significant differences (Table 1).

Fig. 2
figure 2

Flow diagram of enrolment and randomisation of patients

Table 1 Characteristics of patients at baseline

Primary outcome

The primary endpoint occurred five times in four (7.8%) patients in the observation group and 29 times in 18 (33.3%) patients in the control group (p = 0.001), with an overall incidence rate of 16.8 versus 88.1 per 1000 ventilator days, in the observation versus control groups, respectively (p < 0.001). Lip PI occurred in 0 versus 14 (25.9%) patients, with an incidence rate of 0 versus 63.8 per 1000 ventilator days in the observation and control groups, respectively (p < 0.001). Both oral–mucosal and facial PIs showed no significant differences between the two groups (Table 2).

Table 2 Clinical outcomes

Secondary outcomes

No differences were observed between the two groups in terms of the duration of mechanical ventilation, VAP, or tube dislodgement (Table 2).

Discussion

In this study, the observation group had a significantly lower incidence of EIRPI compared to the control group. The incidence of lip PI was significantly reduced, but there were no significant differences in oral–mucosal or facial PIs.

Recent studies have reported different rates of PIs with traditional endotracheal tube fixation devices. Sun et al. reported that the incidence of facial and lip PIs was 58.1% [8]. Qin et al. reported that the incidence of EIRPI was 23.7% and that the lip was the most commonly affected area (76.7%) [9]. In our study, the incidence of EIRPI with a traditional device was 33.3%, which is comparable to the findings of the aforementioned two studies. We hypothesised that the variation in reported incidence may be due to differences in headband tension, as there were no devices to measure and tailor it for each patient.

The National Pressure Ulcer Advisory Panel recognises the vulnerability of mucosal tissue to PIs caused by medical devices such as endotracheal tube fixation devices, emphasising the potential for tissue ulcers due to applied pressure [10]. In this study, the incidence of EIRPI was 8.7% in the observation group and 33.3% in the control group. The results confirmed that the novel device had significant advantages in preventing EIRPI, particularly lip PI.

We conclude the reasons for this to be:

  • Suitability of the contact area: The contact of the novel device with the skin of the face compared to the lips, it is less prone to mechanical damage. The facial skin has a higher degree of keratinisation, which provides stronger protective function [11].

  • Differences in the environment at the contact areas: The complex environment of the lips, including factors such as oral secretions and bacteria, increases the risk of PI. The moist environment of the lips weakens the mucosal barrier function [12] and predisposes to secondary infection. On the other hand, the environment of the facial skin is simpler, and dressings placed at the skin–silicone interfaces can prevent the accumulation of skin moisture, reducing the occurrence of PI [13].

  • The use of silicone pads and dressings in the novel device promotes reduced and uniform pressure between the skin and device, thereby reducing the occurrence of PI [14,15,16].

Overall, the advantages of the novel device include shifting the point of force application away from the lips prone to EIRPI, minimising the risk of mechanical damage to the lips.

In our previous study, the novel device was used to successfully treat lip PI in a 79-year-old patient with diabetes [17].

This study has several limitations, including a dual-centre design, small sample size, limited research duration, and imprecise control of headband tension, and the control group is not a standard of care in every ICU. Moreover, the open-label study design with a subjective outcome, i.e. the number of PIs, could introduce bias into the findings. Future research should aim to improve methodology by integrating more objective assessment tools.

To improve understanding of the novel device used in mechanically ventilated patients, multi-centre studies are needed. Future research should investigate incorporating sensors into endotracheal tube fixation devices to monitor real-time data on skin conditions, blood perfusion, and tissue cell deformation, potentially reducing the incidence of EIRPI.

Conclusions

This study demonstrated that the novel device has therapeutic potential in preventing EIRPI, especially lip PI.

Availability of data and materials

The datasets generated and/or analysed during the current study are available from the corresponding author on reasonable request.

Abbreviations

EIRPI:

Endotracheal intubation-related pressure injury

PI:

Pressure injury

ICU:

Intensive care unit

VAP:

Ventilator-associated pneumonia

References

  1. Qian Y, Lu H. Design and clinical application effect analysis of a new type of oral fluid suction device for patients with orotracheal intubation. Contrast Media Mol Imaging. 2022;2022:6057115.

    Article  PubMed  PubMed Central  Google Scholar 

  2. Martin M, Decamps P, Seguin A, Garret C, Crosby L, Zambon O, et al. Nationwide survey on training and device utilization during tracheal intubation in French intensive care units. Ann Intensive Care. 2020;10(1):2.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  3. Amrani G, Gefen A. Which endotracheal tube location minimises the device-related pressure ulcer risk: The centre or a corner of the mouth. Int Wound J. 2020;17(2):268–76.

    Article  PubMed  Google Scholar 

  4. Shimizu T, Mizutani T, Yamashita S, Hagiya K, Tanaka M. Endotracheal tube extubation force: adhesive tape versus endotracheal tube holder. Respir Care. 2011;56(11):1825–9.

    Article  PubMed  Google Scholar 

  5. Kayser SA, VanGilder CA, Ayello EA, Lachenbruch C. Prevalence and analysis of medical device-related pressure injuries: results from the international pressure ulcer prevalence survey. Adv Skin Wound Care. 2018;31(6):276–85.

    Article  PubMed  PubMed Central  Google Scholar 

  6. Rashvand F, Shamekhi L, Rafiei H, Nosrataghaei M. Incidence and risk factors for medical device-related pressure ulcers: the first report in this regard in Iran. Int Wound J. 2020;17(2):436–42.

    Article  PubMed  Google Scholar 

  7. Zhang Y, Wang S, Zhang X, Zhang W, Wang X. Incidence and influencing factors of medical adhesive-related skin injury in critically Ill patients. Adv Skin Wound Care. 2020;33(5):260–6.

    Article  PubMed  Google Scholar 

  8. Sun Y, Fan H, Song XX, Zhang H. Comparison of three fixation methods for orotracheal intubation in 95 adults. Eur J Med Res. 2020;25(1):45.

    Article  PubMed  PubMed Central  Google Scholar 

  9. Qin L, Yun W, Hang C. Risk factors of endotracheal intubation-related pressure injury among patients admitted to the ICU. Adv Skin Wound Care. 2021;34(3):144–8.

    Article  PubMed  Google Scholar 

  10. National Pressure Ulcer Advisory Panel. Mucosal pressure ulcers: an NPUAP position statement. Washington, DC; 2008 Aug. http://www.npuap.org/wp-content/uploads/2012/01/ Mucosal_Pressure_Ulcer_Position_Statement_final.pdf

  11. Delmore BA, Ayello EA. CE: pressure injuries caused by medical devices and other objects: a clinical update. Am J Nurs. 2017;117(12):36–45.

    Article  PubMed  Google Scholar 

  12. Hajhosseini B, Longaker MT, Gurtner GC. Pressure injury. Ann Surg. 2020;271(4):671–9.

    Article  PubMed  Google Scholar 

  13. Orlov A, Gefen A. Differences in prophylactic performance across wound dressing types used to protect from device-related pressure ulcers caused by a continuous positive airway pressure mask. Int Wound J. 2023;20(4):942–60.

    Article  PubMed  Google Scholar 

  14. Arundel L, Irani E, Barkema G. Reducing the incidence of medical device-related pressure injuries from use of cpap/bipap masks: a quality improvement project. J Wound Ostomy Continence Nurs. 2021;48(2):108–14.

    Article  PubMed  Google Scholar 

  15. Hughes MA. Silicone pressure-reducing pads for the prevention and treatment of pressure ulcers. Br J Community Nurs. 2014;19(Suppl6):S46-52.

    Article  Google Scholar 

  16. European pressure ulcer advisory Panel (EPUAP), national pressure injury advisoryPanel (NPlAP), Pan pacific pressure injury alliance (PPPlA). In: Prevention andtreatment of pressure ulcers/injuries: clinical practice guideline. Australia:Osborne Park: 2019

  17. Zhang Q, Zhang X, Han C, You J, Zhao Y, Xu R, et al. Successful treatment of endotracheal intubation-related lip pressure injury using a self-developed fixation device. Crit Care. 2023;27(1):298.

    Article  PubMed  PubMed Central  Google Scholar 

Download references

Acknowledgements

The authors would like to thank the medical staff in the intensive care unit and pressure injury specialist care team at Nanjing Lishui People’s Hospital and Xuzhou City Central Hospital for their assistance with the trial.

Funding

Supported in part, by Research and Teaching Special Project of Hubei University of Science and Technology (Project No.: 2023YJDJY15), by Nanjing Health Science and Technology Development Special Fund Project (Project No.: YKK23241), and by the National Science and Technology Major Project of China (2022YFC2504505 and 2022YFC2504400).

Author information

Authors and Affiliations

Authors

Contributions

XDZ and SQL led the process and drafted the manuscript. RX, BTY, and YG edited the manuscript. QBZ and JQY collected all data. QBZ, JQY, and YG performed all statistical analyses. BTY and YG drafted the tables and figures. XDZ, QBZ, JQY, RX, ZJZ, YLS, CRH, SYZ, BTY, YG, and SQL critically revised the manuscript. All authors read and approved the final version of the manuscript.

Corresponding authors

Correspondence to Bangtao Yao, Yan Geng or Songqiao Liu.

Ethics declarations

Ethics approval and consent to participate

The study was performed in accordance with the Declaration of Helsinki and was approved by the Institutional Review Board of Nanjing Lishui People’s Hospital and the Medical Ethics Committee of Xuzhou City Central Hospital. All participants provided their written informed consent to participate in this study.

Consent for publication

The authors declare to have informed consent of the patient to publish the image.

Competing interests

The authors declare that they have no competing interests.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Zhang, X., Zhang, Q., You, J. et al. Effect of a self-developed fixation device on preventing endotracheal intubation-related pressure injury: a randomised controlled trial. Crit Care 28, 87 (2024). https://0-doi-org.brum.beds.ac.uk/10.1186/s13054-024-04874-7